Our full technical support staff does not monitor this forum. If you need assistance from a member of our staff, please submit your question from the Ask a Question page.


Log in or register to post/reply in the forum.

CR1000Xe not running with CR1000X program


TLT May 7, 2025 01:38 PM

Hi,

We cannot get our first set of CR1000Xe running with the same program that works in a CR1000X. We tried replaced a CR1000X to the same set up and load it with the same program and it works fine. We can confirm the both AVW200 and 3 set of multiplexers are in working condition. We suspect that the CR1000Xe did not open the serial port connected the AVW200. We are getting '0' for all frequency readings.

The both AVW200 are power by the 12V and a SW12 to power the 3 set of multiplexers.

Do anyone know what caused this?

Please help. Thanks in advance.

'Declare Public Variables

Public PTemp, batt_volt

Public DST(96,6)

Public VWHz(96)

Public AVW_Result(4)

'Declare Other Variables

Dim x()


'''Define VW Data Tables

DataTable (LOGGER_VW,1,-1) 

Sample (1,batt_volt,FP2) 

Sample (1,PTemp,FP2) 

Sample (96,VWHz(),IEEE4) 

Sample (3,AVW_Result(),FP2)

EndTable


'Main Program

BeginProg

 SW12 (SW12_1,1 )

 Scan (3,Min,0,0)

PanelTemp (PTemp,250)

Battery (batt_volt) 

'''Start of MUX #1 and #2     

SerialOpen (ComC1,38400,0,0,10000) 'Turn on ComC1 (C1-C2) for direct communication with AVW200.    

AVW200 (AVW_Result(1),ComC1,0,200,DST(1,1),1,1,32,Start_Freq_Mux1,End_Freq_Mux1,2,_60Hz,1,0)

For x = 1 To 32 

VWHz(x) = DST(x,1)

 Next x


 AVW200 (AVW_Result(2),ComC1,0,200,DST(33,1),2,1,32,Start_Freq_Mux2,End_Freq_Mux2,2,_60Hz,1,0)

 For x = 33 To 64    

VWHz(x) = DST(x,1)             

Next x

 '''Start of MUX #3

SerialOpen (ComC3,38400,0,0,10000) 'Turn on ComC3 (C3-C4) for direct communication with AVW200.  

AVW200 (AVW_Result(3),ComC3,0,201,DST(65,1),1,1,32,Start_Freq_Mux3,End_Freq_Mux3,2,_60Hz,1,0)

For x = 65 To 96    

VWHz(x) = DST(x,1)

Next x

CallTable LOGGER_VW 'Store a set of readings

NextScan '''End Scan  

EndProg

 

Thanks,

TLT


TLT May 7, 2025 03:02 PM

Hi,

I post it again because not sure why the first post went under review after edited it.

We cannot get our first set of CR1000Xe running with CR1000X program.

Everything works fine after we replaced the CR1000Xe to a CR1000X into the set up. So we can confirm the both AVW200 and the 3 set of multplexers are in good condition. We suspect the CR1000Xe did not open the serial port to communicate with the AVW200, there is no clicking sound coming from the multiplexers. All frequency readings are '0'.

At the beginning , all the AVW200 and the multiplexers were power by a single 12V on CR1000Xe. Then we thought could be the power not enough to power them, so we switch the 3 set of multiplexers to a SW12, but it still does not work.

We cannot find what is the cause. Does anyone know what is the fault that caused this issue?

Please help.

Thanks in advance.

'Declare Public Variables

Public PTemp, batt_volt

Public DST(96,6)

Public VWHz(96)

Public AVW_Result(4)



'Declare Other Variables

Dim x()



'''Define VW Data Tables

DataTable (LOGGER_VW,1,-1) 

Sample (1,batt_volt,FP2) 

Sample (1,PTemp,FP2) 

Sample (96,VWHz(),IEEE4) 

Sample (3,AVW_Result(),FP2)

EndTable


'Main Program

BeginProg    

SW12 (SW12_1,1 )  

Scan (3,Min,0,0)

PanelTemp (PTemp,250)

Battery (batt_volt)    

'''Start of MUX #1 and #2     

SerialOpen (ComC1,38400,0,0,10000) 'Turn on ComC1 (C1-C2) for direct communication with AVW200.    

AVW200 (AVW_Result(1),ComC1,0,200,DST(1,1),1,1,32,Start_Freq_Mux1,End_Freq_Mux1,2,_60Hz,1,0)

For x = 1 To 32    

VWHz(x) = DST(x,1)               

Next x
   

AVW200 (AVW_Result(2),ComC1,0,200,DST(33,1),2,1,32,Start_Freq_Mux2,End_Freq_Mux2,2,_60Hz,1,0)

For x = 33 To 64    

VWHz(x) = DST(x,1)             

Next x 

'''Start of MUX #3

SerialOpen (ComC3,38400,0,0,10000) 'Turn on ComC3 (C3-C4) for direct communication with AVW200.        

AVW200 (AVW_Result(3),ComC3,0,201,DST(65,1),1,1,32,Start_Freq_Mux3,End_Freq_Mux3,2,_60Hz,1,0)

For x = 65 To 96    

VWHz(x) = DST(x,1)             

Next x
 

CallTable LOGGER_VW 'Store a set of readings 

 NextScan '''End Scan  

EndProg 

Thanks,

TLT


JDavis May 7, 2025 10:29 PM

If PakbusEncryption is enabled on the data logger in settings, that would prevent talking to the AVW200 modules unless you add EncryptExempt() to the program.


TLT May 8, 2025 02:12 PM

Hi good day JDavis,

Thanks for your great help.

Problem solved.

Thanks,

TLT

Log in or register to post/reply in the forum.